Attenuation in soft tissue is closest to:

Master the Sonography Principles and Instrumentation Exam. Hone your skills with flashcards and detailed multiple choice questions. Prepare confidently with real exam scenarios and in-depth explanations. Ready yourself for success!

Multiple Choice

Attenuation in soft tissue is closest to:

Explanation:
Attenuation in soft tissue increases with frequency and is described as an attenuation coefficient in dB per centimeter per MHz. For soft tissue, a typical value is about 0.5–0.7 dB/cm/MHz, meaning every centimeter of tissue at 1 MHz would attenuate roughly 0.5–0.7 dB, and higher frequencies produce proportionally more attenuation. The figure closest to this standard range is 0.6 dB/cm/MHz, so it best fits what soft tissue does in practice. Other options either miss the per-MHz scaling or give numbers that don’t align with how soft-tissue attenuation behaves across clinical frequencies.

Attenuation in soft tissue increases with frequency and is described as an attenuation coefficient in dB per centimeter per MHz. For soft tissue, a typical value is about 0.5–0.7 dB/cm/MHz, meaning every centimeter of tissue at 1 MHz would attenuate roughly 0.5–0.7 dB, and higher frequencies produce proportionally more attenuation. The figure closest to this standard range is 0.6 dB/cm/MHz, so it best fits what soft tissue does in practice. Other options either miss the per-MHz scaling or give numbers that don’t align with how soft-tissue attenuation behaves across clinical frequencies.